Stimulus rate question

Stimulus rate question

Postby Masud Asadi » Fri Jun 01, 2012 7:40 am

Dear Stephan
Hello :I am very glad to contact directly with Otodynamic experts
I am an Iranian Audiologist and I use ILO288 with ILOV6 software . one issue to ask is Rate of signal This parameter cannot be modified My question is that why rate is slow and remain unchanged
Thank in advance

[Note from moderator: This topic has been moved from the General section]
Masud Asadi
 
Posts: 1
Joined: Fri Apr 06, 2012 6:53 pm

Re: Hello

Postby Stephen Casey » Fri Jun 08, 2012 12:19 pm

Hello Masud,

Thank you for your query, and sorry for the delayed reply. I have passed your question on to a colleague who will post a reply very soon. We may start a new thread for it, so that it's easier for other people to find.

Thanks again.

Steve.
Last edited by Stephen Casey on Fri Jun 08, 2012 12:26 pm, edited 3 times in total.
Reason: Too many 'query'(s)!
Stephen Casey
 
Posts: 4
Joined: Tue Mar 27, 2012 11:30 am
Location: United Kingdom

Re: Hello

Postby Oliver » Mon Jun 11, 2012 9:00 am

Dear Masud,

The stimulus rate for TEOAE is dependent on the length of the response window that is required. Unlike Auditory Brainstem Responses the stimulus rate does not significantly alter the response recorded within the response window, so the stimulation rate is set to the maximum possible for the response window set.
The response window is set on V6 in Options->Stims->Stim type. The 'Linear' and 'Non linear' settings have response window of 20.48ms - giving a stimulus rate of 48.8Hz. The 'Quickscreen' setting has a response window of 12.8ms giving a stimulus rate of 78.1Hz.
Low frequency TEOAEs are recorded later in the response window than high frequency OAEs, so the 'Quickscreen' 12.8ms window will reduce the amplitude of low frequency emissions recorded, but will collect data faster.
You describe the presentation rate as 'slow' this may be due to the way the NLo figure is calculated; the figure on V6 'NLo' represents sets of 8 (for 'Non Linear') or 16 (for 'Quickscreen') stimuli presented, not individual stimuli.

I hope this answers your questions,

Yours,

Oliver Brill
Product Scientist
Otodynamics